Signal Conditioning
SCXI 1000 Tc module failure of channel #8 multiple failures.different systems.
Several of my test systems have had a SCXI 1102 module channel #8 failure. WHen repaired it was found to be a bad amplifier on that channel. Why?
$600 to fix a bad component.
Is there a built in failure mode to this module SCXI1102?
Is there an internal issue causing this failure?
(3,291 Views)
There is no trend that we have seen to cause such an error.
Can I contact you directly to dicsucss this and resolve the reason for the damage to your equipment?
Anuj D.
(3,282 Views)
Thanks for the reply, 920.683.7855 is my phone on the floor. These systems have a SCXI 1000 chassis, SCXI1102 in slot 1, SCXI1102 in slot 2, SCXI1102C in slot 3. Tc all have grounded cables, to a TBX1303 terminal board to CJC the Tc outside the test room. the SCXI1102C is connected to a PTB power monitor board, several triggering devices (CT with rectifier to signal (Vdc) coil energized or not), and pressure transducers with a .1 to 5.1 Vdc input/output. a regulated 24 Vdc is used to supply the voltage to the transducers. could a spike in the supply 24 Vdc voltage smoke the amplifier on the SCXI1102 , even though they are isolated?
please give me a call with any insite you can give to the issue is appreciated
Scottyd
(3,280 Views)