Signal Conditioning
Stairstepping in thermocouple data with TC-2095/SCXI 1102c
Here's my setup:
32 K-type thermocouples (mixture of grounded sheathed and ungrounded exposed tip)
TC-2095 rack mount thermocouple connector block
SCXI 1102c module
I am having an issue of stairstepping in the thermocouple data I am getting.  The system seems to be sampling at the 500-1000 Hz sampling rate because noise and gentle sloping is recorded in the tc data in between the major jumps in temperature.  The stairsteps generally occur at 1 or 2 Hz.  The only thing I modified on the hardware was swapping in the 10 Mohm resistor packs into the tc-2095 to prevent a grounding issue from ruining the readings from our grounded thermocouples.  Any thoughts on how to fix the stairstepping?
Alex
(6,125 Views)
Hey,
Did this work without the10Mohm resistor pack?? Did this ever work?
You say you have grounded thermocouples---what are they grounded to?  Can you expand on this setup.
How did you configure the dip switches on the TC-2095 for each thermocouple? I'm referencing the settings mentioned in table1 of the tc-2095 specifications sheet.  This is found on page 5-6 of thespecs manual: http://www.ni.com/pdf/manuals/321731b.pdf
Can you attach a screenshot of the data that you are seeing so that I can inspect it.  Also what sampling rates are you specifying?  how are you programing this (Labwindows/labview?? etc... )
Please post back with this info so we can work to resolve this issue thank u
Charley Dahan
Global Account Manager
(6,103 Views)