 _natalie_
			_natalie_
			10-22-2019 08:39 AM
I am having difficulties controlling a PXI-2567. I am using this relay driver with a PCIe-1073 chassis and PCIe-8361. It took a very long time to be able to communicate with the PXI-2567. After pulling the module out of the slot and putting it back in, I was able to communicate with it as show in the NI-MAX screen shot below.
Using a VI I found under examples (Help --> Find Examples --> NI Switch --> Switch - Controlling an Individual Relay(NI-Switch).vi), I closed k8 as shown in the screen show below.
If I go back to NI Max, click on the PXI-2567 and open test panel, the NI-Switch Front Panel Opens. If I click Relays and under "Relay Information", I can see that k8's relay position is closed.
However, if I connect a DMM to measure continuity between ch8 and com8 (pin 1 and pin 3), it appears open as shown below (the red lead is floating).
Am I missing something, or is it possible that there is something wrong with relay driver module? Any help is greatly appreciated.
Thanks.
10-22-2019 08:59 AM
Natalie,
I assume that you did, but please confirm that you powered off the chassis, removed and added the module back in, then powered on the chassis and rebooted. Is that correct?
Do any of the other relay connections work? Did you try only one bank or both banks?
Looking at the image it seems like one LED (the 5V LED) is not powered on. This might indicate that a fuse is blown. If the fuse is blown the relays may not function as expected.
10-22-2019 09:19 AM
Hi
@Frank-L wrote:
Natalie,
Hi Frank,
Thanks for getting back to me. Yes, the chassis was definitely powered off when I took the module in and out. I have tried other relay banks.
Is there a way to turn on and off the internal power supplies? Or should both LEDs just remain lit if the fuses aren't blown? Since we just got this module the other week and have not really used it yet, it would be surprising that a fuse is blown, but you never know.
10-22-2019 10:14 AM
Natalie,
These power supplies are exposed on the front panel, there is a chance that something was shorted? Unfortunately, there is no way to turn off the internal power supplies.
Both LEDs should remain lit if the fuses aren't blown. You can disconnect the board and run a continuity test on the fuses to see if the fuses have been blown.http://www.ni.com/pdf/manuals/373469e.pdf
Perhaps work with customer support to fix the board if the module is new and within warranty?
10-22-2019 10:42 AM
Thanks again. I will look at this later today and get back to you.
We will not be using the internal power supplies. However, if I understand correctly, if a fuses is broken, the module will not work properly. Is that correct?
10-22-2019 11:38 AM
I checked the fuses and neither are damaged. Could it be something else on the board?
 HarrisonKb
			HarrisonKb
			11-09-2020 10:25 AM
Has there been any update on this? I just started using some old 2567 boards at my facility and I have been having the same issue. The relays will work fine for a few days, then I will have the same issue where I open/close the relay on the NI, but the module appears to remain open. What else can I do to troubleshoot this issue?
11-09-2020 11:32 AM
It turns out my module was defective, and I was sent a new one.
If I understand you correctly, you are able to open/close a relay in NI Max, but it physically is not.  Have you tried using one of the other slots with your module?
 HarrisonKb
			HarrisonKb
			11-10-2020 07:36 AM
Thanks for the follow-up. I've diagnosed my issue a little further. As other pins have been working fine, I believe the heat shrink on my solders to the 78-pin D-sub failed and cased some shorts between pins. I've purchased a terminal block board to connect to the 2567 and I will see if that helps.