04-07-2008 10:13 AM - edited 04-07-2008 10:19 AM
04-14-2008 08:37 AM
Hello Chris,
The only way to know for sure is by using GPIB analyzers to see the handshaking, control, and data line transitions. Since communication works fine when the isolator is not on the bus, I would start with the isolator. The best test would be to use the bridge on the other side of isolator and use an analyzer on each side. This way you can monitor the line transitions on each side. The isolator should propogate all signals as the come in (except serial polls which are usually a special case).
You will either find that the isolator or the instrument is misbehaving by not putting the lines in the proper state. Use your NI-488.2 manual to find out which node should be controlling eah of the GPIB lines.
Controller - IFC, ATN
Talker - DAV
Listener - NRFD, NDAC
Steven T.
04-16-2008 02:25 AM - edited 04-16-2008 02:25 AM
05-06-2008 08:15 AM - edited 05-06-2008 08:16 AM
05-07-2008 11:14 AM
Chris,
Thats good news! Its also alot easier of a fix than tracing analyzer captures for a few days 🙂
Steven T.