02-13-2012 10:45 AM
I will change this in the program. Thank you for the help
02-13-2012 10:48 AM
Change what? There is nothing wrong as far as I can tell. Actually put a probe on the VISA Write input first and report what is there.
02-17-2012 04:40 PM
I was trying to apply the changes that Jeff mentioned. However I couldn't get it to work. So I am still not able to turn off the auto range on my device. The auto range is causing delayed measurements in my VI.
02-20-2012 09:55 AM
Well, did you do what I asked? Did you actually probe what the VISA Write is sending? Did you turn on NI-Spy/NI-Trace to capture the write? Do one or the other and compare what the manual says to turn ofauto-ranging.
02-21-2012 11:58 AM
I just edited the read output vi and it took care of my problem. thanks for the help
02-21-2012 01:02 PM
Attach what you have done. I don't understand how a modification to a read can change what you said what a problem writing.
02-21-2012 03:02 PM
Here it is.
10-21-2012 03:55 PM
Can you post a screenshot of the pre-/after changes? I am using the supplied VI's as well and the configure measurement function just hangs on me. What i mean is that the Over Current flashes on the instrument, but it does not overcome this.
Thank You
10-22-2012 04:31 PM
Hi GHernandez,
It sounds like you're getting a different problem than what was previously posted. Are you using the drivers available on this page?
Cheers,
10-22-2012 07:02 PM
Yes. I am using the VIs supplied by National Instruments. I've attached screenshots of the subVI calls and the Main Vi. I made a test vi that will call this function. It hangs at setting the current limit. It will flash on the front panel and not do anything else.