02-12-2014 12:17 PM
hi i have an agilent B2901A current source and im trying to out put 100V... since its capable to about 200V but i can go abouve 21V. When i enter anything about 21V it only sends out 21V. Anyine know how to fix this. Is it my compliance values or range?
Thanks in advance
Solved! Go to Solution.
02-12-2014 12:24 PM
What current is it putting out @ 21V with your circuit?
Cameron
02-12-2014 12:51 PM - edited 02-12-2014 12:54 PM
about 21mA... its set up with a aproximate 1kohm resistor
02-12-2014 01:14 PM
Are you using two channels? Looking at the data sheet for the instrument it appears you can't use more than 21V if using 2 channels. I haven't looked at the manual for it so I could be wrong.
02-12-2014 01:24 PM
@adgurrol wrote:
about 21mA... its set up with a aproximate 1kohm resistor
If you are set up as a constant current source, then you should be putting out 21V (21mA*1kOhm = 21V). Looks like it is working exactly as it was designed.
If you truely want 100V out, then you need to change it to a constant voltage source and increase your compliance current to at least 100mA.
02-19-2014 12:04 PM
no just 1 channel.
02-19-2014 12:08 PM
i believe i do have it set up as a constant voltage source and even when i do it manually and not through labview i can not seem to go above 21V even witht the complaince surrent at 100mA or higher?
02-19-2014 12:10 PM
02-19-2014 12:57 PM
so i got it so i can go higher manually but i cant go as high through labview as i was manually... the higherst i got manually was 42V with 100mA limit
02-19-2014 01:14 PM
ok nvm i figured it out thanks for all your help it turns out if i want to go above 42V i need an interlock terminal. the problem through labview was mu limit compliance was too high.