how to get RMS of triangle wave
The image you posted of your data looks suspiciously like it is seriously undersampled.  It is definitely not a triangle wave.  The plot just draws straight lines between data points, making the image look "pointy."  Does this data make any sense when compared to what you expect the signal source to generate?
Try sampling 10 or 100 times faster to see what the waveform looks like.
Any calculations you do on this data are likely to be rather meaningless if it is sampled at less that the Nyquist rate.
Lynn
Message 11 of 22
(9,588 Views)
Johnsold is right.
Assuming your main signal frequence is line freq. (50/60Hz) you should at least measure with 2kHz better 10kHz to make a reasonable 'True' RMS measurement.
Make a measurement with the highest possible samplerate (250kHz?) and have a look at your signal (or use a scope) . Even better : Make that reading and take a look at the  FFT power spectrum of that waveform. All that points different from zero in the power spectrum diagram represent energie for your RMS value. Most propably the main content is in the lower range and multiple of the line freq. (And post that diagram 😉 )
You might find an example in the provided examples where all this is already done....
EDIT: And capture 5 to 10 periodes ....
Greetings from Germany
Henrik
LV since v3.1
Message 12 of 22
(9,581 Views)
Hi Jill, thanks for your input.  The image I posted was sampled at 500us.  By separating the cycles, taking the peak of each cycle, then averaging the peaks and dividing by sq.rt(3) I get my expected voltage.  Now I'm confused, how is this possible if it is not a triangle wave?  I sampled the signal again at 1us which can be seen in the attached photo.  Unfortunately, I cannot sample any faster using my cRIO.
Message 13 of 22
(9,570 Views)
OK your math is a bit funny.  I don't follow the Vpk *1/3^-2  This would estimate a Veff of a sine wave.
Vrms is the area under the waveform
For a pure sine wave Vrms = 0.658 Vpk and Veff = .707 Vpk
For a pure square wave Vrms = Vpk
For a pure Triangle wave Vrms = 0.5Vpk
Message 14 of 22
(9,556 Views)
The RMS value for a triangle wave (or sawtooth) is Vpk/sqrt(3).  For symmetric waveforms like those you only need to integrate over a quarter period to verify (or use the LV functions).  At any rate, by inspection it must be less than the sinusoid Vpk/sqrt(2).
How well this simple estimate matches the calculated value depends on the input (1) being triangular or sawtooth (2) having zero dc offset (3) being sufficiently sampled so that the simple numerical integration is accurate and (4) having a sample period equal to an integer number of full periods of the squared waveform.
Message 15 of 22
(9,547 Views)
Accepted by phdieum
OK.  Forget rms and simple formulas.  The graph image you posted sampled at 500 us shows that your signal is quite noisy and not clearly either a sine or triangle as the underlying structure.  It appears that the noise peak to peak amplitude is about 1/4 of the signal peak to peak amplitude at best.
Let's go back to the beginning.
1. What is the source of this signal?
2. What is the source or sources of the noise?
3. How do the frequency and amplitude of signal and noise vary?
4. What are your going to do with the data when you get it? Or, Why do you think rms is the appropriate measurement technique for what you are doing?
5. Do you have other knowledge which can help analyze the signal, such as a clean synchronizing signal?
Lynn
Message 16 of 22
(9,536 Views)
EDIT=  double checking my math
Message 17 of 22
(9,534 Views)
Yup For a sine wave
But sin2ωt + cos2ωt = 1. Therefore the average values of either of them must be 1/2.
Therefore the rms value of Iosinωt must be Io /√ 2
for a Triangle (AC, with 0 DC offset) A =1/2(H*B).  Integrate over B = 1/2 Vpk.
Message 18 of 22
(9,527 Views)
If you do not believe me, do you believe G?
Keep in mind that the mean of the square is typically not the square of the mean so the RMS is not equivalent to the area.
Message 19 of 22
(9,525 Views)
If you are using a 9206 the max samplerate is 250kS/s (4µ) or did you use another card?
Do you have access to a scope? If you don't know your signal all will be just guessing ( And don't trust the Klein CL2000 DMM , it's only valid for frequencies up to 400Hz )
Henrik
LV since v3.1
Message 20 of 22
(9,509 Views)