Hi everyone!

 

I was wondering what the parameter accuracy of the NI-6008 means. It is 14.7 mv in RSE configuration. 

 

Does this number show the minimum value that the Ni-6008 can detect? i.e. the resolution?

 

Thanks

 

Lander

0 Kudos
Message 1 of 7
(3,791 Views)

Accuracy is more of how close the measurement is from the true.  Resolution is the smallest change that can be detected.  While related, they are separate.


GCentral
There are only two ways to tell somebody thanks: Kudos and Marked Solutions
Unofficial Forum Rules and Guidelines
"Not that we are sufficient in ourselves to claim anything as coming from us, but our sufficiency is from God" - 2 Corinthians 3:5
0 Kudos
Message 2 of 7
(3,789 Views)

Got it. Thanks!!

 

So the 14.7mV would be the absolute accuracy and the resolution would be 10-(-10)/2^11 =9.76 mV. I do not find this value anywhere.

 

Thanks again.

 

Lander

0 Kudos
Message 3 of 7
(3,778 Views)

Apparently they left the resolution for you to calculate, which you did.  Note also that the noise can be 5 Vrms.  It is possible that the accuracy figure was based on averages which eliminated the noise effects.  In a real world system you might see 2 or 3 LSB variations due to noise and accuracy limitations.

 

Lynn

0 Kudos
Message 4 of 7
(3,771 Views)

You calculation should be (10- -10)/(2^12) = 4.88mV.  So the resolution at the full range is 4.88mV.  An accuracy of 14mV makes more sense with this.


GCentral
There are only two ways to tell somebody thanks: Kudos and Marked Solutions
Unofficial Forum Rules and Guidelines
"Not that we are sufficient in ourselves to claim anything as coming from us, but our sufficiency is from God" - 2 Corinthians 3:5
0 Kudos
Message 5 of 7
(3,766 Views)

The specified resolution for the USB-6008/6009 is one bit less when using a single ended configuration. So 2^11 is the correct factor.

 

Lynn

 

Resolution 6008.png

0 Kudos
Message 6 of 7
(3,761 Views)

OK! Thank you all very much!