09-07-2012 08:49 AM
Hi everyone!
I was wondering what the parameter accuracy of the NI-6008 means. It is 14.7 mv in RSE configuration.
Does this number show the minimum value that the Ni-6008 can detect? i.e. the resolution?
Thanks
Lander
09-07-2012 10:13 AM
Accuracy is more of how close the measurement is from the true. Resolution is the smallest change that can be detected. While related, they are separate.
09-08-2012 03:16 AM
Got it. Thanks!!
So the 14.7mV would be the absolute accuracy and the resolution would be 10-(-10)/2^11 =9.76 mV. I do not find this value anywhere.
Thanks again.
Lander
09-08-2012 10:27 AM
Apparently they left the resolution for you to calculate, which you did. Note also that the noise can be 5 Vrms. It is possible that the accuracy figure was based on averages which eliminated the noise effects. In a real world system you might see 2 or 3 LSB variations due to noise and accuracy limitations.
Lynn
09-08-2012 01:06 PM
You calculation should be (10- -10)/(2^12) = 4.88mV. So the resolution at the full range is 4.88mV. An accuracy of 14mV makes more sense with this.
09-08-2012 03:00 PM
The specified resolution for the USB-6008/6009 is one bit less when using a single ended configuration. So 2^11 is the correct factor.
Lynn
09-10-2012 01:39 AM
OK! Thank you all very much!