03-30-2007 09:56 AM
Andrew,
I am running LabView 7.1. This VI was created in ver 8.2 and won't run on my system.
Where do the drivers come from to talk to the PXI cards? Are they part of Cal Exec? Is it possible that I have an out of date or missing driver? I believe that Ni-DAQmx ver 8.3.1 is installed as well.
Thanks
Richard
03-30-2007 11:22 AM
Hi Richard,
I've attached the 7.1 version of the VI. The only driver you should need for the 4461 is the DAQmx driver, and 8.3.1 should be fine. Let me know how it goes.
Andrew S
03-30-2007 02:27 PM
I get some errors when trying to run the VI stating that some sub and poly VI's are broken or not executable.
How do you include screen captures or attach files in this forum?
03-30-2007 02:43 PM - edited 03-30-2007 02:43 PM
Message Edited by stilly32 on 03-30-2007 02:43 PM
08-09-2012 07:37 AM
Hi Andrew,
Do you have any idea regarding this error?
This is in case of PXI-4472?
Thanks!
--
Łukasz
08-10-2012 01:52 PM
Hi lukko,
I'm assuming this self-calibrate is being done in MAX, correct? Does this device pass the self-test? Does this error occur on multiple cards? Have you tried searching for any bent pins, updating the DAQmx drivers, changing slots or changing chassis?
08-27-2012 04:55 AM
Hi Carisa,
Thanks for the reply.
That is correct, I use self-calibrate feature in MAX. I have actually 3 PXI-4472 boards bought at the same time but only one of those cannot go through the self-calibrate procedure. The board can pass the self-test and is capable of doing acquisition. I also tried to change the slots as well as the chassis. I use the newest (I think) DAQmx in version 9.5.1f1. Still I have the same Error-200717.
In general the systematic error is not so high and for my application is enough however I would like to have everything working correctly. Please let me know if you have any ideas.
Best regards,
--
Łukasz
08-28-2012 08:27 AM
Hi Lukasz,
Thanks so much for answering my questions. After some further research it seems like there is an internal hardware issue (potentially a fuse) and the card should be sent in for repair in order to clear the 200717 error. Hope this helps answer your question.
Thanks,
Carisa L.
02-04-2013 03:16 AM
Hello, Carisa and colleagues!
Have some problem for PXI-4461 (with PXI-8101 controller + PXI-1042 chassis) :
in MAX into PXI-4461 tab doing:
1) self-test: ok!
2) self-calibration: error -200717
Measurement taken during calibration produced an invalid AI offset calibration constant.
If performing an external calibration, ensure that .........................................
Carisa and colleagues, how can clear this error without " ni technical center"?
Thank you:)
02-04-2013 07:01 PM
Hi Andrey_Z,
In all cases like this that I've seen, it was necessary to send in the card for repair. There is not a way to clear it persay.
Thanks,
Carisa L.