01-03-2006 01:43 PM
01-05-2006 11:43 AM
01-05-2006 12:10 PM
Thank your for your reply.
Are you sure that this can be done with my configuration? I have 2 SCXI-1127 cards and here's what I was told:
"There is no way to do handshake scanning with this hardware setup. What you would need is the PXI-1011 chassis.
These 1127 can only send its Advance Complete trigger to SCXI trigger “TTL 2”, available only in the PXI-1011. It cannot send their Advance Complete trigger the rear connector."Thank you
Jerry.
01-05-2006
01:23 PM
- last edited on
05-15-2024
09:12 AM
by
Content Cleaner
You are correct, I had missed this detail of the SCXI-1127's operation. Since you are using a PCI-4070, it is not possible to use the PXI-1011 chassis with your current DMM. The best solution I could recommend would be to use another device, such as a USB-1359 to control the SCXI chassis, then the AUX Connector on your DMM will be available to make the trigger connections needed for handshaking with the SCXI-1127 switch. Keep in mind; however, that you will only be able to handshake with one of your switches using this method. I believe that you are already working directly with another Engineer at NI, so I will allow him to assist you from this point forward.
NI Products: USB-1359
Jason S.
Applications Engineer
National Instruments