SignalExpress
Tektronix TDS2014B. The scope free runs just fine under it's own control but under SignalExpress control it repeatedly triggers creating a recording that is discontinuous bursts of data.
I am using a Tektronix TDS2014B.  The scope free runs just fine under it's own control but under SignalExpress control it repeatedly triggers creating a recording that is discontinuous bursts of data. The scope's memory is not sufficient for the duration of some tests that I need to run, so, I was hoping SignalExpress was the answer.
By the way, this is a recently installed instance of the SignalExpress software running in 30 day trial mode.
The story is the same as found in the thread "http://forums.ni.com/t5/SignalExpress/Using-SignalExpress-for-data-logging-on-oscilloscope-Tektronix...".
Message 1 of 53
(8,293 Views)
In the "horizontal" settings tab, what are all the options you can select for "Acquisition Mode"?  The other setting of interest is the "Type" on the "Trigger" tab.
Unfortunately I don't have the hardware in order to test this, and I can't find it documented in the SignalExpress Tektronics Users Manual.
Systems Engineer
SISU
Message 2 of 53
(8,281 Views)
Please see the attached pdf for setup information. Let me know if there are more specifics that will help solve this.
Message 3 of 53
(8,277 Views)
I don't have the hardware so I can't try this, but you might try changing the "Acquisition mode" to something other than "Sample".  Is there a "Continuous" Opton?
Systems Engineer
SISU
Message 4 of 53
(8,248 Views)
Unfortunately, no continuous option. If only I could fool the system with a trigger condition that would always be met,
Message 5 of 53
(8,241 Views)
What are all the different trigger types you can select on the "Triggers" tab?
Systems Engineer
SISU
Message 6 of 53
(8,193 Views)
Trigger tab has these options:
Type - Edge; Video or Pulse
My present choices of other settings under the trigger tab are:
Source - For this portion of my test I am using only channels 3 and 4 and have tried triggering from either one
Coupling - DC
Level - 0.00V (have tried various other small levels with the same result
Slope - Rising
Holdoff - 500ns
To ensure there were no conflicts, I uninstalled the Tektronics OpenChoice software and the NI software as well. Then, I reinstalled the NI package. I found a threead that mentioned a .dll file for the USB connection could be a problem if OpenChoice was installed after LabVIEW SignalExpress is installed. Still trigger, stop, trigger, stop, trigger, stop............... and so on.
Message 7 of 53
(8,189 Views)
Another thing to try is to use the generic IVI drivers.  So setup your step to use a IVI Scope Acquire instead.  I simulated a scope and used IVI drivers and was able to get continuous measurements.  (In Signal Express 2012).
Systems Engineer
SISU
Message 8 of 53
(7,970 Views)
What is the firmware version on your Scope?  The firmware is the code on the device that determines how the instrument responds to different commands.
Also, which version of Signal Express are you using?
Systems Engineer
SISU
Message 9 of 53
(7,903 Views)
Haven't had a chance to try generic IVIs yet. Regarding the scope firmware, it is listed in the Utilities / Misc. screen as follows:
Misc.
TDS2014B
Version FV:v23.08
Serial Number C101067
GBIP Setup         Address: 2
TEK-USB-488
      None
The NI SignalExpress is Version 2.5.1 (2007 vintage, however, this was recently downloaded from the Tektronics site).
I am beginning to wonder if the scope's USB interface can handle the data rate produced by free running. I have tried even the slowest horizontal rate to minimize the sample rate (to no avail, though).
Message 10 of 53
(7,897 Views)